Dear all,
thank you very much for your input.
I'll have to experiment a bit more with the distance to the walls (which probably contain metal) and the chair.
Thank you also for the idea with the second sensor, I hadn't tried this before.
Best,
Julian
Am 05.06.2013 um 17:53 schrieb Elizabeth Anne Bock, Ms:
> Hi Julian,
> We have experienced this problem as well. We solved it using the following guidelines:
>
> No metal near the polhemus or any of the receivers/transmitters - you will need to move the setup around the room to find the perfect spot.
> Use a wooden or plastic chair
> Use plastic or cloth glasses/holder to attach the receiver to the subject
>
> My system is sensitive to the proximity of the transmitter and the receiver. I use two receivers, #1 is the stylus and #2 is secured to plastic glasses that the subject wears. The transmitter is taped to the back of the chair. If #2 and the transmitter are too close to each other (i.e. a short person or child), then the measurement are inaccurate. You would have to experiment with different distances that give good results.
>
> Hope this helps!
> Beth
>
> ------------------------------------------------------------------------------------------
> Elizabeth Bock / MEG System Engineer
> McConnell Brain Imaging Centre / Montreal Neurological Institute
> McGill University / 3801 University St. / Montreal, QC H3A 2B4
>
> Office: 514.398.3706
> MEG Lab: 514.398.6056
> Mobile: 514.718.6342
> From: fieldtrip-bounces at science.ru.nl [fieldtrip-bounces at science.ru.nl] on behalf of Julian Keil [julian.keil at gmail.com]
> Sent: Monday, June 03, 2013 11:14 AM
> To: FieldTrip discussion list
> Subject: [FieldTrip] Polhemus Patriot
>
> Dear FieldTrip-Users,
>
> I have a not really FieldTrip-related question, but maybe one of you can help me anyways.
> In our lab, we have a Polhemus Patriot 3D tracking system to acquire electrode positions.
> Unfortunately, the recordings are severely distorted in the Z-axis (up-down).
> After contacting the Polhemus Support, I got the information, that this is due to metal in the surroundings of the source which distorts the magnetic field.
> I tried to get as far away from any metal as possible in our lab (~1.5 m to the walls and floor) but to no avail.
>
> Now to my question: Has anyone any experience dealing with this? I'm quite puzzled by this as I know plenty of labs use Polhemus trackers, and I'm not sure if our lab is especially metal prone or if I'm missing something.
>
> Thanks a lot for any help.
>
> Julian
>
> ********************
> Dr. Julian Keil
>
> AG Multisensorische Integration
> Psychiatrische Universitätsklinik
> der Charité im St. Hedwig-Krankenhaus
> Große Hamburger Straße 5-11, Raum E 307
> 10115 Berlin
>
> Telefon: +49-30-2311-1879
> Fax: +49-30-2311-2209
> http://psy-ccm.charite.de/forschung/bildgebung/ag_multisensorische_integration
>
> _______________________________________________
> fieldtrip mailing list
> fieldtrip at donders.ru.nl
> http://mailman.science.ru.nl/mailman/listinfo/fieldtrip
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://mailman.science.ru.nl/pipermail/fieldtrip/attachments/20130605/1eb74243/attachment-0002.html>
More information about the fieldtrip
mailing list