    Hi Daria,<br>
    <br>
    Also, did you try looking at your raw data (so before the ERPs are
    made) in FieldTrip and in Net Station? So open your .raw data in
    NetStation. You can just look at a random participant's data, and
    check if the scale is the same. I'm pretty sure no amplification
    takes place when reading in the data into FieldTrip, this check can
    confirm that. Hope this helps, and let us know what you find.<br>
    <br>
    Cheers,<br>
    Ingrid<br>
    <br>
    <br>
    <div class="moz-cite-prefix">On 9/3/2012 1:35 PM, Robert Oostenveld
      wrote:<br>
    </div>
    <blockquote
      cite="mid:63F6ADC4-BA8E-4299-973F-3DDAFE630C51@donders.ru.nl"
      type="cite">Hi Daria,
      <div><br>
      </div>
      <div>There might indeed be an incompatibility in the gain  value
        applied to the data.</div>
      <div><br>
      </div>
      <div>If you are able to export the data using the EGI software in
        another format (e.g. EDF), you could read them both into
        FieldTrip and compare the timecourse aplitudes to check whether
        they are different. In case they are, I suggest to involve EGI
        support. They are aware of the FieldTrip reading functions and
        will know much more about the internals of their various file
        formats.</div>
      <div><br>
      </div>
      <div>best regards</div>
      <div>Robert</div>
      <div><br>
      </div>
      <div><br>
      </div>
      <div><br>
      </div>
      <div><br>
        <div>
          <div>On 3 Sep 2012, at 17:22, Daria Laptinskaya wrote:</div>
          <br class="Apple-interchange-newline">
          <blockquote type="cite">Dear Fieldtrippers,<br>
            <br>
            I analysed EGI-data (.raw-files, NetStation 4.3.) with the
            fieldtrip toolbox. <br>
            The ERPs look perfect. <br>
            The 'only' problem is, that my signals look to strong.<br>
            It looks like if I get amplitutes (average) until 70 <span>µ</span>V.
            <br>
            I think it must be an intern amplifier for the signals.<br>
            <p class="MsoNormal"><span>Does anyone know this problem
                with EGI-data?</span></p>
            <p class="MsoNormal">Or does anyone know, where I can see
              the amplifier?<br>
            </p>
            <p class="MsoNormal"><br>
              <span></span></p>
            <p class="MsoNormal"><span><br>
              </span></p>
            <p class="MsoNormal"><span>Greetings,</span></p>
            <span>Daria</span>
            _______________________________________________<br>
            fieldtrip mailing list<br>
            <a moz-do-not-send="true"
              href="mailto:fieldtrip@donders.ru.nl">fieldtrip@donders.ru.nl</a><br>
            <a class="moz-txt-link-freetext" href="http://mailman.science.ru.nl/mailman/listinfo/fieldtrip">http://mailman.science.ru.nl/mailman/listinfo/fieldtrip</a></blockquote>
        </div>
        <br>
      </div>
      <br>
      <fieldset class="mimeAttachmentHeader"></fieldset>
      <br>
      <pre wrap="">_______________________________________________
fieldtrip mailing list
<a class="moz-txt-link-abbreviated" href="mailto:fieldtrip@donders.ru.nl">fieldtrip@donders.ru.nl</a>
<a class="moz-txt-link-freetext" href="http://mailman.science.ru.nl/mailman/listinfo/fieldtrip">http://mailman.science.ru.nl/mailman/listinfo/fieldtrip</a></pre>
    </blockquote>
    <br>
    <pre class="moz-signature" cols="72">-- 
Ingrid Nieuwenhuis PhD
Postdoctoral Fellow
Sleep and Neuroimaging Laboratory
Department of Psychology
University of California, Berkeley
California 94720-1650
Tolman Hall, room 5305</pre>
  </body>
</html>